I am trying to determine the applied pesticide concentration in artificially coated seeds. Have tried various methods unsuccessfully. Cannot get complete removal of the coating or complete dissolution into organic solvent. Anyone with experience in this field?Any help will be much appreciated
It would help us to know the composition of the seed coating, the name of the pesticide and which techniques have you tried?
Don Shelly
LGC Standards
Hi

The seed coating consists of a polymer (most likely polyvinyl alcohol), mica and pigment. The details are a bit hazy, because the companies are afraid to reveal the exact components of the coating. Currently looking for clothianidin, imidacloprid and abamectin.

I have done some solvent solubility tests. Most of the coating seems to be water soluble. I have tried lowering the pH to increase solubility and this seems to be working. Acetone seems to extract the pigment from the coating, but there is quite a lot of the solid coating left on the seeds. Methanol does not work at all. There is limited dissolution with acetonitrile.

The water appears to dissolve/swell the polymer. Next step ought to be solvent extraction. Or perhaps a modified quechers extraction so that I can extract the pesticides and transfer to organic solvent suited to GCMS analysis. (Except for abamectin, which I will have to remove via clean-up [thinking SPE]).

Thanking you for any suggestions
You might try dissolving in warm water with agitation. Extract and clean-up using a buffered QuEChERS. Use matrix matched calibration.
Don Shelly
LGC Standards
Great idea. Thanks
What about a high-speed homogenizer? That should pulverize everything, allow the solvent to reach the analyte.
Thank you for all your suggestions. I swelled the polymer with warm water and agitation. Then used a quechers extraction. In the process of validation now. Getting beautiful fairly clean chromatograms
fatskhan wrote:
Thank you for all your suggestions. I swelled the polymer with warm water and agitation. Then used a quechers extraction. In the process of validation now. Getting beautiful fairly clean chromatograms



Glad it's working for you. Thanks for the feedback.

Don
Don Shelly
LGC Standards
8 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry