Hi,
My name is Melody and I'm new in this area. I been working in sample preparation since January of this year because for my thesis I have to quantify PAH's in sediment and soil. I read a lot of articles for extraction methods and the one that was most suitable for the supplies and equipment of my lab was ultrasound, nevertheless I been having a lot of trouble with this, and my recovery doesn't exceed 60%. Hope you can help me.

I did some preliminar studies and I found the following:
solvent: acetone
time: 20min
vol of solvent: 15ml
cicles: 3 times the above.

I use a liquid cromatographer HPLC with FD for the reading. My ultrasound bath have 35kHz and 285W. I collect the 3 extracts and concentrated by rotaevaporation to 1ml, and then I put it in a 5ml ballon with mobile phase.

Hope you can help me and tell me what am I doing wrong or what can I improve. Thanx in advance.

Melody
Your extraction solvent could be too polar. What is your mobile phase? What is your sample size?
Don Shelly
Don Shelly Consulting, LLC
don.shelly@donshellyconsulting.com
Melody.CO,

First, I have to agree with Don that I suspect acetone is too polar. Second, if you are using an US bath, I found these to be woefully inefficient extraction devices. I would turn to a shaker table before I would turn to an ultrasonic bath. If you are talking a horn, all bets are off.

Best regards,

AICMM
Thanx for the replies. :D

My mobile phase is Acentonitrile:water 75:25 and my sample size is 5 g. I did some test with Hexane:Acetone but it was similar with acetone alone. Some people tell me is my Ultrasound bath that have to low power or frequency. But I don't know.

Thanx a lot.

Melody
Try mixing your soil sample with sodium sulfate anhydrous until it forms a free flowing powder. That should increase surface area and permeability. Use the 1:1 acetone/hexane. Add 1 ml of acetonitrile as a keeper prior to solvent exchange.
Don Shelly
Don Shelly Consulting, LLC
don.shelly@donshellyconsulting.com
Don't know anything about HPLC, but we use methylene chloride for GCMS (We used to use different ratios of MeCl2:acetone but recently switched to 100% MeCl2). Then we solvent exchange into hexane for the cleanup prior to analysis.

As Don stated, we also mix the soils with sodium sulfate.
6 posts Page 1 of 1

Who is online

In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry