i am running a Perkin Elmer balanced pressure static headspace sampler, going to an Agilent GC split/splitless inlet. The split ratio is set to 1:1 and the inlet set to constant pressure at 4.8 psi.
Usually, we have the head pressure set to give a 30mL/min flow out of the split vent. however, we recently got a new HS-40 and an engineer told us that if we reduced the head pressure it should improve the sensitivity of the instrument because we would be losing less out of the split vent. had the pressure set at 18.6 psi originally giving 30mL/min out of the split vent. we reduced this (over a few stages) down to 10psi, which gave 11mL/min out of the split vent.
this made no difference to the sensitivity of the instrument, which has, to be honest, baffled my head a bit. can anyone suggest a reason why changing the head pressure has no effect on the chromatography?
not particularly vital to my work, but just something that has been bugging me!
many thanks,
Beth