I'm setting up a method for measuring methane in seawater samples. I have an SRI GC/FID. Column is 2 m of 1/8" SS packed with HayeSep A, carrier is UHP N2 at 20 mL/min. Oven temp is 50 C. I have made some preliminary injections of air and methane standards using a 1 mL sample loop and have nice symmetrical CH4 peaks, retention time is 0.97 min.
This setup will eventually linked to an equilibrator for sea surface measurements and continuously cycling. I'm expecting that water vapour and other C compounds in seawater will eventually cause a problem with ghost peaks and baseline deterioration in a single column config. so I'm considering installing a pre column with backflush to vent to stop those compounds entering the main column. I'm looking for advice on whether this is the best way go and what would be a good adsorbent for the pre-column. I have been very impressed with a two column system in a Peak Labs H2 analyser under continuous operation (injections of equilibrated headspace). I think that they used Unibeads in the pre column and water vapour didn't seem to be a problem.
Thanks