Hi, I would like to know if anyone habitually uses solid-phase extraction in trace analysis of pesticides.
do you think it is a valid quantitative method of extraction?
despite all the precautions I only get low recoveries (<50%).
do you think this technique can be used for concentrations of order of magnitude of ppt?

Thanks
We use the solid phase extraction listed in EPA Method 525.3 for several pesticides at concentrations in the low to sub ppb range and normally get 70%+ recovery. 70% is the lowest recovery allowed by the method and we usually pass that QC at well above that.

One thing that helps is using matrix matching for the calibration standards. We extract the 1L volume of DI Water through the cartridge and process it through elution by solvents and concentration to final volume, with no surrogate or internal standards added, then use that final solvent to dilute our standards with when preparing them in the normal fashion. The small amount of background matrix compensates for any ionization enhancement or suppression in the MS source.

Also, if you are using GC/ECD EPA Method 508.1 has a very similar extraction process.
The past is there to guide us into the future, not to dwell in.
What type of pesticides? Some of the more polar organo-phosphates will require a different phase than C18.
I'm interesting in organochlorine pesticides. Thanks for the suggestion of matrix matching, I'm tryng.

Greetings
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry