Hi. I have a problem with an acetylcysteine ​​solution. I found an impalpable black precipitate. a third-party laboratory investigation found traces of bismuth. do you know a simple technique to identify bismuth?
thank you
Off the top of my head, atomic absorption spectroscopy. A Google search should turn up lots of references.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Dissolve the ppt. in minimum amount of acid. Adjust pH to 8 or higher with sodium carbonate solution. It isn't bismuth if you don't get a white ppt..

Please describe your problem. Are you preparing an a-cys solution from contaminated a-cys? Its not impalpable.

Bismuth quantitaton requires ICP, AA, AE, IC, CE or other instrumental methods unless you have special chemicals for qualitative spot tests. Chromatographic techniques are preferred, here, such as IC, maybe CE, as well.

_____________

Hi. I have a problem with an acetylcysteine ​​solution. I found an impalpable black precipitate. a third-party laboratory investigation found traces of bismuth. do you know a simple technique to identify bismuth?
thank you
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Majestic-12 [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry