The four standards were first dissolved in water 1:40 (w/v), then mixed 1:1:1:1 (v/v/v/v), filtered (0.45 µm) and finally diluted 1:1 (v/v) with acetonitrile to achieve a 12.5 mg/mL stock solution

Shouldn't the stock solution above be 3.125 mg/ml?

Each of the four standards are different molecules
chemist23 wrote:
The four standards were first dissolved in water 1:40 (w/v), then mixed 1:1:1:1 (v/v/v/v), filtered (0.45 µm) and finally diluted 1:1 (v/v) with acetonitrile to achieve a 12.5 mg/mL stock solution

Shouldn't the stock solution above be 3.125 mg/ml?

Each of the four standards are different molecules


1:40, then 1:4(one standard to 4 total volume) then 1:1 (one standard to 2 total) so you have 40x4x2 dilution factor which is a 1:320 dilution from the original. If the original weight was 1g then you have 1g\320 = 0.003125g/ml

1:40 = 1g/40ml = 0.025g/ml then dilute to 4ml = 0.00625g/ml then dilute in half with acetonitrile = 0.003125g/ml

Worked through two different ways for the same answer so it seems correct at 3.125mg/ml
The past is there to guide us into the future, not to dwell in.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry