Those of you doing SPE extractions. I've been thinking of using glass graduated centrifuge tubes for doing final blow down. Is there a "class A" requirement to get a consistent volume?
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Probably need to check the methods to see if required. We are required by SOP to use them. We use the 1ml Class A volumetric test tubes for our final volume adjustment.
With a graduated centrifuge tube you will at least know you are in the 500-700ul range before transferring. I have gotten pretty good at estimating how much is in the 1ml then using a 1ml syringe to wash the initial tube used for concentration with the amount needed to top it off then using a disposable pipette wash the sides a couple times then use that to bring to volume. If you can get accurate centrifuge tubes then that would probably work even better.
With a graduated centrifuge tube you will at least know you are in the 500-700ul range before transferring. I have gotten pretty good at estimating how much is in the 1ml then using a 1ml syringe to wash the initial tube used for concentration with the amount needed to top it off then using a disposable pipette wash the sides a couple times then use that to bring to volume. If you can get accurate centrifuge tubes then that would probably work even better.
The past is there to guide us into the future, not to dwell in.
2 posts
Page 1 of 1
Who is online
In total there are 11 users online :: 1 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm
Users browsing this forum: Semrush [Bot] and 10 guests
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm
Users browsing this forum: Semrush [Bot] and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science