Is there a reason why the preparation for Class 2 mixture C is not included in the procedure? I am wondering if this standard is necessary for residual solvents analysis as per USP <467>?
-
- Posts: 12
- Joined: Wed Oct 23, 2019 2:31 pm
- Location: UK
I don't know why it isn't a part of <467>.
if your testing as per <467> following procedure A and then B. You only should need the class 1 mixture and the class 2 mixtures A and B.
if your testing as per <467> following procedure A and then B. You only should need the class 1 mixture and the class 2 mixtures A and B.
-
- Posts: 1
- Joined: Tue May 12, 2020 7:21 am
a cumulative procedure may be used to calculate the residual solvent levels in the ... Residual Solvents Class 2—Mixture C RS to a 100-mL volumetric flask, dilute ... tellpopeyes caused by a substance and that disappear after exposure to the substance.
3 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science