Currently I am having a problem with C18 column. In running progress retention time and theoretical plate of the peak decrease gradually (first time retention of peak is 14.4 min, second time is 13.0 min and third time is 11.2 min) and the next analysis after first time analysis some impurities co-eluted. May be the column have losing performance.
I tried many ways to wash this column from water/MeOH/Acetonitrile to acetic acid 1%/MeOH/Acetonitrile. But they are not effective.
So that, now we only run this column with this mobile phase only 2 times. Please help me give the flushing method for wash and restore this column after run this mobile phase.
I used it for analysis Risedronate impurities with mobile phase (I referred from "Risedronate tablets" of BP2020) and program bellow:
Chromatographic condition for Risedronate impurity:
- Column: C18 (150x4.6-mm; 3-µm), Ecosil SH/, PN: AS1203-1546
- Detector: UV 263 nm
- Temperature: 40oC
- Flow rate: 1.0 mL/min
- Injection volume: 20 µL
Buffer solution pH 7.0: Dissolve 1.75 g dibasic potassium phosphate (K2HPO4—174.18) and 0.40 g edetate disodium (C10H14N2O8Na2·2H2O—372.2), 1.7 g tetrabutylammonium hydrogen sulfate in water sufficient 1000 mL. Adjust pH 7.0 by sodium hydroxide 1 N.
Gradient:
Time (minutes)/ % ACN/ % buffer pH 7.0
0 7 93
14 7 93
15 25 75
50 25 75
51 7 93
72 7 93
Diluent: Water
Thank you very much.