Hi there. I´m having trouble with this analysis, amlodipine besylate.
I cannot get concordance within standards and the biggest problem, areas whitin days vary. I´ve tried all hplc we have (Promeinence, 1260 and 1100) and always same problem. One day in an hplc you get some area, and the other day, same hplc, same mobile phase (or new), and response changes. Tried another standard from different source but nothing different happened, same problem. Instruments are qualified, is the only product I´m seeing this. But not with Valsartan, only amlodipine, and the system is the same because it´s a gradient, you quantify both peaks.
Anyone has experience with this monography?
Q. F. Ignacio Viera
Hello. May amlodipine be just unstable?
Best regards,
Dmitriy A. Perlow
You need to work systematically here!

1. Areas change: How? larger/smaller? Progressively or randomly?

2. Do you work wth the sandard or the samples?

3. Do you use the same solutions each day or do you dissolve new samples/standards every day and se these area variations.

Best Regards
Learn Innovate and Share

Dancho Dikov
Thank you guys. It is not an stability problem.
We work with secondary standards, solutions prepared at the moment of use.
Areas differ randomly.
Q. F. Ignacio Viera
4 posts Page 1 of 1

Who is online

In total there are 4 users online :: 0 registered, 0 hidden and 4 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 4 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry