Some time we have seen a small peak in blank with the same retention at the interest peak. Although area of this peak is small, but I wonder that how much % area of this peak compare to the interest peak that we can say it would not impact, and it pass the system suit. How can we calculate it? Any advice, would greatly appreciate.

Best regards,
Maily
Per ICH Q3 a related substance can be no more than 0.05% wt/wt of the API i the drug product. Assuming this peak is a treated like a related substance the tolerable peak area is a maximum of 0.05% of the peak area for the API assay peak.
You can try to move this peak by adjusting the mobile phase composition so R>1.5
HPLC chemist wrote:
Per ICH Q3 a related substance can be no more than 0.05% wt/wt of the API i the drug product. Assuming this peak is a treated like a related substance the tolerable peak area is a maximum of 0.05% of the peak area for the API assay peak.




Thank you so much. I have thought the same.

Maily
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry