Hi all,

I've been tasked with maintaining a rather old agilent 1100 HPLC stack containing the following components:
  • Degasser
  • Bin pump
  • Autosampler
  • VWD

This is connected to a Finnegan LCQ Deca and the entire set of instruments are controlled via Xcalibur.

Recently, the bin pump module began throwing all sorts of pressure errors that weren't easily solvable. We happened to have an identical spare bin pump unit, which I swapped in to the stack. Everything seems to be working now with the pump, but when I attempt to run methods through the Xcalibur software, the system hangs while waiting for contact closure and then crashes. If I manually program the LC using the gameboy control unit, everything seems to run fine. Is this possibly a control software issue?

I'm much more familiar with FPLC systems than HPLC, so I'm a bit stick with how best to diagnose this.
When you re-configured the replacement BIN pump, you forgot to 'activate' the ECI board in the pump module as part of the new configuration. Depending on the software version, the default is often disabled in some systems. You need to check the 'enabled box.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry