Which column is recommended when separating neutral or other highly retained compounds?
you don't know this now but there is something that needs to be said.
That depends on what the "highly retained" compounds are. In the absence of any other details, a good starting point for most people is a good C18 column based on "type B" silica with a "full range" gradient from water (or buffer) to organic solvent.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
tom jupille wrote:
That depends on what the "highly retained" compounds are. In the absence of any other details, a good starting point for most people is a good C18 column based on "type B" silica with a "full range" gradient from water (or buffer) to organic solvent.

Thanks, very helpful!:)
you don't know this now but there is something that needs to be said.
You might want to consider low loading C18, or something with a shorter chain(C4, C8) or more polar stationary phase. This will shorten your method and save some ACN (or MeOH) you will use in the mobile phase.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Dear Rosi,

"Highly Retained" I unerstand as Late Retention Peaks when You using Reverse Phase HPLC columns.

Plese try with HILIC columns f.e Silica Si column and 90% Acetonitrile as mobile eluent HPLC phase. In this way, the higly retained compounds will be eluated first, near the chromatogram front. The larger size column are better because easer packed by manufacter.

Serum clinical sample is depronized with acetonitrill or HCLO4 and next acetonitril.

For example - Ganciclovir is separated on C18 column but highly retained pek appear with rt=22 minute. However, in HILIC mode and Si column, the late pik going in front and ganciclovir posses nice clear chromatogram.
Kazimierz
I doubt that "Rosi" is still reading here, since she only wrote this one post and it is 3 months old, but for the sake of discussion:

We do not know, if she just wanted to get rid of the late eluters, or if these were her analytes.

I would second the advice to try a simple C18. One could change the strong eluent from MeOH or ACN to IPA/ACN. We use this eluent for HPLC of intact triglycerides - works well and the backpressure is comparable to MeOH.

HILIC will lead, as Kazimierz wrote, to the disappearance of the RP-late-eluters in the solvent front. I once had the task of developing a method, where very hydrophobic substances (glycolipid biosurfactants) had to be separated in HILIC mode. I ended up using ACN with 40% DCM and 10% water in ACN as eluents. I don't know if this can still be called HILIC, but it worked - even though there was no water in the weak eluent.
6 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot], Google Feedfetcher and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry