Validation of HPLC Methods

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hii Everyone,
We have developed and validated assay and related substances methods for a drug substances. This methods have different conditions compared to those used at the ASMF holder. Authorities want to that cross validation data for assay and related substances of three drug substance batches should be provided.
The validation study performed for the assay and related substance method .The parameters described here are according to ICH guidelines and include accuracy, precision, specificity, limit of quantitation, linearity, range, solution stability and robustness.
Do I have to perform cross validation?
Is there any guideline about cross validation? How can I explain that cross validation isn't necessary?


Many Thanks
Please,can anyone answer my question?

Many thanks
Essentially, you have to do this to show that your method is equivalent to the compendial or "referee" method. There is no transitive property of validation for analytical methods.
Thanks,
DR
Image
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry