Posted: Tue Jun 11, 2019 5:05 pm
Hi, I have been doing optimizing a new HPLC method on a new "aqueous" column and ran into some older issues that I had on a spherisORB ODS2 C18 column. There is a persistent blob that always comes off the column.
At around 4 min past the solvent front we begin to see the weird blob. Here are some things I have tried to fix the problem.
HPLC- columbic cell detector
- replace the mobile and flush column
- clean and replace rheodyne parts
- replace sample loop
- try different percentages of MeOH (does seem to move blob slightly)
- injecting pure MeOH makes the blob really big
When I just do the trigger I don't see the blob, however, if I inject mobile phase into the loop I see the blob. If I inject water I see the blob or if I inject the sample I also see the blob.
At around 4 min past the solvent front we begin to see the weird blob. Here are some things I have tried to fix the problem.
HPLC- columbic cell detector
- replace the mobile and flush column
- clean and replace rheodyne parts
- replace sample loop
- try different percentages of MeOH (does seem to move blob slightly)
- injecting pure MeOH makes the blob really big
When I just do the trigger I don't see the blob, however, if I inject mobile phase into the loop I see the blob. If I inject water I see the blob or if I inject the sample I also see the blob.