lmh » Mon Jun 10, 2019 9:04 am
If you are using a high-pressure binary system with a column fitted between the pump and the mixer, to remove a contaminant from the solvent, then once the clean-up column has become fully equilibrated with solvent, it will not affect the retention times.
Although I have personally never found it necessary, I know of many reputable people who have used a clean-up column in this way, once they've traced the cause of a ghost peak to a contaminant in their solvent.
Extra volume before the mixer should not affect the gradient. That is the thinking behind binary systems compared to quaternary.
If the tubing becomes so kinked that it develops a leak, then yes, there will be drastic differences in the elution, because you'll then have a lower flow, and the wrong percentage, and possibly even weird dead-volume issues from the mixer itself if, at any point during the gradient, the pump on the leaky side is running at such a low flow that there is a net flow from mixer to leak.
I am also assuming the mixer is the right one for the job, and that the system now is in its 'bad' state... the most difficult situation is if the system now is in its 'good' state, and unfortunately your method isn't working, while the method used to work, for some obscure reason connected with a fault (such as a small leak on the clean-up column that no one had noticed) that you've now corrected.
I like your suggestions of a bit more priming and a few conditioning runs - certainly can't do any harm, and might sort out a lot of the more trivial causes of unreproducible retention times. Check the solvent really is what you think it is, and any buffers have been made correctly. What does the system pressure look like now, compared to when the method worked? From resolution of 2.5 to coelution is quite a drastic loss of quality - something moderately drastic must have happened to cause it!