Mobile Phase in (HILIC)?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hello Everyone,

Recently we bought Atlantis, HILIC OBD prep 5 µm, Dimesion 19 mm× 150mm column for the purification of hydrophilic compounds.

The mobile phase mostly used in HILIC are 90/10 acetonitrile/ 100mM ammonium formate pH3.

My question is it necessary to use buffer ammonium formate or acetate to retain hydrophilic compounds, as my compound are sensitive to low pH.

Can I used only water and Acetonitrile to achieve good separation?
any suggestion will be highly appreciated.

Regards.
If you compound is neutral or surface of silica does not have ionizable groups of opposite charge to your analyte you can avoid buffer. If your compound has ionizable groups which can interact with stationary phase by ion-exchange mechanism then you need ions in the mobile phase to facilitate elution.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Basic analytes are the most polar (or at least the best-retained in HILIC). This widespread use of buffers with pH of 3 or so reflects applications involving peptides. Below pH 4, peptides lose the (-) charge on carboxyl- groups. This leaves them with a net (+) charge. They are retained most strongly under those conditions. If your analyte of interest doesn't have any carboxyl- groups or if increasing retention time isn't a concern (i.e., being able to get adequate retention at a lower % ACN), then there's no need to use so low a pH. Use ammonium formate or acetate right out of the bottle if you like (with resulting pH's around 5.5 and 6.5, respectively). I do recommend having some salt present, if only to insure that there's a reproducible layer of counterions on top of the (presumably somewhat charged) stationary phase surface. That will affect reproducibility. Try it with and without 20 mM salt (that's 20 mM overall, not just in the aqueous portion of the mobile phase).
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
Vlad Orlovsky wrote:
If you compound is neutral or surface of silica does not have ionizable groups of opposite charge to your analyte you can avoid buffer. If your compound has ionizable groups which can interact with stationary phase by ion-exchange mechanism then you need ions in the mobile phase to facilitate elution.
Organic_chemist wrote:
Hello Everyone,

Recently we bought Atlantis, HILIC OBD prep 5 µm, Dimesion 19 mm× 150mm column for the purification of hydrophilic compounds.

The mobile phase mostly used in HILIC are 90/10 acetonitrile/ 100mM ammonium formate pH3.

My question is it necessary to use buffer ammonium formate or acetate to retain hydrophilic compounds, as my compound are sensitive to low pH.

Can I used only water and Acetonitrile to achieve good separation?
any suggestion will be highly appreciated.

Regards.



Thanks you for the explanations. It's really helpful
Andy Alpert wrote:
Basic analytes are the most polar (or at least the best-retained in HILIC). This widespread use of buffers with pH of 3 or so reflects applications involving peptides. Below pH 4, peptides lose the (-) charge on carboxyl- groups. This leaves them with a net (+) charge. They are retained most strongly under those conditions. If your analyte of interest doesn't have any carboxyl- groups or if increasing retention time isn't a concern (i.e., being able to get adequate retention at a lower % ACN), then there's no need to use so low a pH. Use ammonium formate or acetate right out of the bottle if you like (with resulting pH's around 5.5 and 6.5, respectively). I do recommend having some salt present, if only to insure that there's a reproducible layer of counterions on top of the (presumably somewhat charged) stationary phase surface. That will affect reproducibility. Try it with and without 20 mM salt (that's 20 mM overall, not just in the aqueous portion of the mobile phase).


Thanks you for the explanation.
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry