Pressure and leak tests Agilent 1100 HPLC
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
Recently I started having an issue with the pressure ripple being unstable. I replaced the piston seals and cleaned the pistons and housing. I also replaced the gold seal on the outlet ball valve and the PTFE frit inside the purge valve, but I'm still having the same issue. I don't see any external leaks, so I'm trying to run the leak test using Chemstation (since I don't have an integrator to plug into the pump). However, the instructions in the manual are unclear on how to initialize the test. I've loaded the leak test method, connected 100% LC-grade IPA to channel D, opened the purge valve and set the flow to 5ml/min for 10 mins, connected the restriction capillary and flushed the pump channel, and then set the flow to 0ml/min and connected the blank nut. However, I'm not seeing where I'm supposed to execute the test. There is not 'execute' button for me to press to begin the test. Is there a sequence that needs to be loaded for this to run? I was having a similar issue with the pressure test. Any help would be greatly appreciated. Thanks!
-
- Posts: 72
- Joined: Sat Mar 28, 2009 12:01 am
Hi Tylerkai,
Agilent moved a lot of the testing and diagnostic functions from Chemstation into their Lab Advisor software at one point. It could be that you need the Lab Advisor software to do these tests. If you don’t have it, I think it can be downloaded from Agilent for free. If not you can find links for it from the Agilent community site
Hope this helps.
Kevin
Agilent moved a lot of the testing and diagnostic functions from Chemstation into their Lab Advisor software at one point. It could be that you need the Lab Advisor software to do these tests. If you don’t have it, I think it can be downloaded from Agilent for free. If not you can find links for it from the Agilent community site
Hope this helps.
Kevin
-
- Posts: 3510
- Joined: Mon Aug 30, 2004 11:43 pm
Before we got the OpenLab stuff with its own inherent "fun", we used A.09 Chemstation. I made I own Methods.M in Chemstation for the two flow/pressure tests.
-
- Posts: 11
- Joined: Tue Aug 28, 2018 3:49 pm
I'll give both of those things a try, thank you!
-
- Posts: 11
- Joined: Tue Aug 28, 2018 3:49 pm
It looks like the pressure and leak tests were moved to the Lab Advisor software. It isn't available for free unfortunately, but I was able to get it and run the test. Looks like I have to replace the active inlet valve since I've already replaced everything else. Thanks for the information!
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science