Akthmps88 wrote:
So, I ordered the Sigma HCG standard (2500 IU per vial). I reconstituted in 1mL of water and injected 40uL onto the HPLC. I used a Waters Symmetry300 C4 column, 5um, 3.9 x 150mm. Column oven is maintained at 40C, flow rate is 1.5 mL/min and mobile phase is as follows:
Mobile Phase A: 0.1% TFA in Water
Mobile Phase B: 0.1% TFA in Acetonitrile
For the first 1 minute, I maintain the gradient at 98% A + 2% B. Then, from 1-25 minutes, it is a linear gradient increasing to 95% B.
At 220nm what I see are a cluster of 4 peaks (with a few smaller peaks trailing off). These peaks present between approximately 8-10 minutes. The biggest of these peaks is around 150 mA tall.
Not sure what to make of these results. Seems it would be hard to do any sort of quantitation with these small, numerous peaks. I was hoping for 2 distinctive peaks, but that is not the case.
Any thoughts on my method setup and/or results?
I'm thinking about taking this method setup and running it to a Charged Aerosol Detector rather than a diode array. Perhaps these subunits do not have a good chromophore?
Thanks for any additional thoughts.
Your method sounds good to me.
I honestly wouldn't bother with Charged Aerosol - if it's a peptide, which it is, it's going to have a rather nice absorbance at 214-220 nm because of the chromophore that peptide bond is.
That being said, I don't have personal experience with CAD.
Check the UV spectra of the peaks that you have, it should help you distinguish which of the peaks are the alpha and beta subunit of HCG.If all have the same spectra, typical for peptides, than what HPLC chemist said about HCG variants could be true.
HPLC chemist wrote:
Try 214 nm where the peptide peak has its maximum absorbance. What is your injection volume? I had to use 250 uL for 9.6 ug/mL (about 10 U/mL) of Oxytocin.
The peaks could be active HCG, deactivated HCG, denatured HCG, deamidated HCG.... You're lucky you only had 4 peaks! I had 6!
214 nm and 0.1% TFA tends to introduce errors into automatic integration in my experience, so I advised him to use 220 nm.
Miniscule difference in sensitivity, much less interference from TFA.
He injected 40 ul to achieve concentration I had advised him and he got 150 mA peak in there, which, going by memory sounds about right on a normal system.