How to seperate close peaks in C18 HPLC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hello everyone.

I am currently trying to seperate compound from an active fraction of a sea star. Based on the nmr i ran earlier, there are sugar and nucleoside compounds in it. I had tried various methods, and this is the method with the best seperation:

This is the link to the chromatogram:
https://drive.google.com/file/d/1jaWg6J ... p=drivesdk

Solvent A: 0.1% TFA in H20
Solvent B: pure Acetonitrile
Flow rate: 1.0 ml/min
Temperature: Ambient
Tyoe of column: Purospher Hibar C18 encapped( 5 micrometer)
Method: 0 to 10th minute: Solvent B from 0% to 25%
10th to 15th minute: Solvent B from 25% to
50%

Can anyone please suggest methods to improve this chromatogram?
I have tried reducing the flow rate to 0.5ml/min and tweaking the gradient, the peaks became wide and separation is not good.

Thabk you in advance.
For me, the easiest thing to try is to make the temperature 10C cooler, try that. Then make 10C warmer than ambient, try that. See if better/worse/same.
for me, some infos are missing:
- dimension of column (diameter x length)
- type of detection (wavelength; do you have PDA or MS at hand?)
- full gradient table (including re-equilibration)
- solvent in which your sample is dissolved
- how does a chromatogram of a solvent blank look like
Those peaks that I see look to be eluting early. The reply above mine has a lot of good questions that would allow us to help solve the problem. Also, please indicate which peaks are of interest to you.
Slowing flow rate is not going to to too much. You compounds elute close to the void and you might run into a problem of interference coming from sample matrix, large injection, column contamination, etc. If you have any ionizable groups in your compound your best choice is mixed-mode chromatography - either HILIC/ion-exchange or RP/ion-exchange. Mixed-mode is all about selectivity and if you can employ two mechanisms you should be able to have retention and separation. Contact me if you have questions.

Here are couple of examples (including nucleotides):
https://helixchrom.com/compounds/guanosine-diphosphate/
https://helixchrom.com/compounds/raffinose/
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry