How to change flow path filter in Merck Hitachi pump L-7100
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
-
- Posts: 511
- Joined: Tue Jul 19, 2011 9:26 pm
Put it in the fridge for a few hours. The Teflon ring will shrink a bit, so usually you can pull the frit out afterwards.
Alternatively, you can try to blow it out with high-pressure air.
Alternatively, you can try to blow it out with high-pressure air.