Acceptace criteria for interlaboratory metod transfer

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hello everyone,
I am looking for some acceptance criteria for interlaboratory non-statistical comparison for related substances method transfer for film coated tablets. Can you, please, advice me? Is there any guideline related to this topic?
Thank You!
1. We did such cGMP transfers to our QC and to contract manufacturers' QC several times a year.

2. Unfortunately, I cannot remember what our agreement values were for agreement to R&D values, or where the values came from.

Obviously, results had to meet cGMP system suitability criteria. This held for EACH OPERATOR at these locations. We supplied three samples previously assayed at R&D; each sample would be prepared from scratch three times, and each prepared sample was injected at least twice.

I think we used 98 – 102% agreement and standard deviation of 2 for each of the nine samples.
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry