Hello,

I'm developing a HPLC method to retain some polar compounds that have weak UV absorption (only one acid/ester group and sometimes an amide group). The route I've chosen to pursue is using a 100% aqueous initial gradient on a suitable column (have tried both the Atlantis T3 and Synergy Hydro-RP) with phosphoric acid as the modifier as formate/acetate kill the sensitivity for these compounds.

For both columns and using either phosphoric acid or 10 mM ammonium phosphate, the issue always persists. It has happened on two HPLCs as well.

If I start at 100% aqueous mobile phase, with no hold time or a 2-5 min hold, the negative peak is there. If I start with 2% or 5% organic, then the peak is massively reduced, though I don't get the retention I need.

The thing that is stumping me is why there's such a sharp peak in gradient blanks (no injection). It seems to be correlated to when acetonitrile starts getting introduced into the gradient.

Relative to 0 mAU at t0, the peak is about the same magnitude in single wavelength chromatograms (~-60 to -70 mAU) across the entire UV range I collected on my DAD (190 - 400 nm). Below is an isoabsorbance plot, spectrum of this peak, and chromatogram at 210nm of a no-injection gradient blank. Note the negative dip (represented by black color):

Image

Method info:
A: 0.1% H3PO4 in HPLC-grade water
B: HPLC-grade acetonitrile

0-4 min: 0%B
4-10 min: 0-80%B
10-12 min: 80%B
12.1 min: 0%B

Phenomenex Hydro-RP, 150 x 4.6mm, 4um, 1 mL/min


Any ideas? Thank you!
dja
What you see is an refractive index issue especially at 210nm. If you run 100% ACN and add water to the mobile phase you would see the same sharp peak, but positive peak.
Gerhard Kratz, Kratz_Gerhard@web.de
2 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry