-
- Posts: 1
- Joined: Thu Aug 27, 2015 7:03 pm
Agilent 1290 Pump Leak Rate Test failure
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
Agilent 1290 Pump Leak Rate Test failure: I have an Agilent 1290 Binary Pump that consistently fails the Pump Leak Rate Test. Testing @1200 bar with 100% IPA. "A" side is fine. "B" secondary passes, and primary starts out OK, but after a few seconds, the pressure drops ~300bar, then goes back up to 1200, leaving a v-shaped notch in the top of the trace. I have tried sonicating the check valves, repeated purging; nothing seems to work. Any suggestions?
-
- Posts: 422
- Joined: Tue Sep 15, 2015 2:07 pm
Sounds like the piston seals may be worn out. Do you visually 'See' a leak anywhere (i.e. MP valve, under pump heads).
-
- Posts: 1
- Joined: Thu Dec 20, 2018 3:26 am
@Multidimensional,
please don't speculate if you are not sure.
@NickW,
first has all the other test leak passed on the Agilent LabAdvisor, if yes then there is no issue. I believe you are running the Leak Rate Test using Agilent LabAdvisor. When performing the Leak Rate Test on Secondary Pump Head there is no issue as the Pressure is able to maintain at 1000Bar. once it has completed that part the Pump will release the pressure and rebuild the pressure in Primary Pump Head. At this moment the Pump should only Pressurize up to 800 Bar. Instead of 800 Bar the Pump actually exceed the Pressure limit and Pressure up to 1200 Bar which cause this issue. This is the a Software Bug. as long as all the other pressure test passes there shouldn't be big issue.
please don't speculate if you are not sure.
@NickW,
first has all the other test leak passed on the Agilent LabAdvisor, if yes then there is no issue. I believe you are running the Leak Rate Test using Agilent LabAdvisor. When performing the Leak Rate Test on Secondary Pump Head there is no issue as the Pressure is able to maintain at 1000Bar. once it has completed that part the Pump will release the pressure and rebuild the pressure in Primary Pump Head. At this moment the Pump should only Pressurize up to 800 Bar. Instead of 800 Bar the Pump actually exceed the Pressure limit and Pressure up to 1200 Bar which cause this issue. This is the a Software Bug. as long as all the other pressure test passes there shouldn't be big issue.
-
- Posts: 422
- Joined: Tue Sep 15, 2015 2:07 pm
Newby, I never speculate. Advice is based on many years of professionally servicing these modules. Without additional information, no clear diagnosis can be made.
4 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science