- Posts: 6
- Joined: Sat Dec 01, 2012 8:12 pm
problem in autosampler shimadzu prominence SIL-20AC
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
hi every one. i was working on shimadzu hplc after analysis was over i kept autosampler for rinsing but after few seconds a sound was produced and autosampler's display showed home pump error any suggestions will be appreciated.
-
- Posts: 158
- Joined: Wed Mar 05, 2008 2:56 pm
- Location: UK
Hi
In the first instance try power cycling it. the needle rinse will home. If that doesn't work remove the power lead, remove the covers and clean the opto sensors. Shimadzus are a bit of a pain for homing due to their sensors.
In the first instance try power cycling it. the needle rinse will home. If that doesn't work remove the power lead, remove the covers and clean the opto sensors. Shimadzus are a bit of a pain for homing due to their sensors.
-
- Posts: 6
- Joined: Sat Dec 01, 2012 8:12 pm
thank you
-
- Posts: 1
- Joined: Wed Sep 21, 2016 11:26 am
hi,
When switched on the HPLC shimadzuu system , initialization failure happening at model number SIL-20A HT / Prominence autosampler with display of SIL ERR PUMP HOME. At the time of initialization, there was strucking sound at measuring pump. Please suggest me the way to troubleshoot this problem.
When switched on the HPLC shimadzuu system , initialization failure happening at model number SIL-20A HT / Prominence autosampler with display of SIL ERR PUMP HOME. At the time of initialization, there was strucking sound at measuring pump. Please suggest me the way to troubleshoot this problem.
-
- Posts: 2
- Joined: Wed Sep 05, 2018 1:38 pm
For those searching ERR HPV HOME: And in case anyone encounters this later... If you disassemble the right side/panel of your SIL 20AC ht, and look up under the HPV; there is a small, thin, green panel with an optical sensor on it... this must be cleared of dust in order for your sample needle to move back "home."
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science