Hey, I have been suffering a ELSD signal unstable problems for a while. Here are the conditions we are using

Purity by HPLC
HPLC System: Agilent 1100/1200 HPLC System
Flow Rate: 0.5 mL/min
Mobile Phase A: 0.04% TFA in Acetonitrile
Mobile Phase B: 0.04% TFA in Water
Column Temperature: 50°C ± 5°C
Autosampler Temperature: Ambient
Run Time: 16 minutes

Detection System: Shimazu ELSD-LT II
Temperature: 75°C
Gas Pressure: 50 psi
Detector Environment: 50HƵ

However, the %RSD for system suitability is out of 10%. Could anyone have an idea about solving the problem. Appreciate it!
What are you analyzing? Make sure that your compound is not volatile, otherwise at lower concentrations you will lose linearity. Try to reduce temperature to 50 or even 35 degrees Celsius.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
To follow up with what was said previously. I think the ELSD temperature is too high. Try nearer 30C to see if that works better.
Kevin
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry