Has anyone done related substances test in 522 EP monograph for pholcodine monohydrate. I am experiencing difficulties in eluting imp G from Reference solution D; I am using EP CRS standard for identification Y0001865. I am using Phenomenex Luna Phenyl Hexyl 2.7u 4.6mm 7.5 cm column. I have absolutely no peaks eluting past pholcodine on this column and according to EDQM I should have 3 peaks with imp G about 4.2 RRT. Diluent and MP I have made as per monograph. Any ideas?

Thank you
Two questions here,

1. Does your chromatogram otherwise match that of the EP ?
2. Is your column brand new, or well used?

Regards,
JMB
Yes, otherwise chromatography matches EP. Retention times a bit shifted but column was not new. I have ordered pristine Kinetex Phenylhexyl and waiting for result.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry