Hi,


I have developed a method for cleaning validation of dexmedetomidine. Concentration is very low (0.0000001 mg / ml=%100 standard). I hava obtained a peak with HPLC and LC MS MS. However I have not provided linearity in both methods. Area of %120 standard less than % 100 standard. I have tried methods too much. Is the range not appropriate in the methods.

Thank you
Good afternoon,

This may help a bit:

https://www.europeanpharmaceuticalrevie ... n-samples/
MattM