Agilent Chemstation Rev. B.04.02 not initializing

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hello,
I am trying to set up an 1100 Series HPLC for one of our labs. I have been able to connect and configure the instrument, however the software will not initialize. The software stops on the Data Analysis portion of the initialization. I have checked the CHEMSTATION.INI to see if the paths for initialization were correct, and they all seem to be. I also went through the Windows OS configuration steps provided by the Agilent Manual, and have uninstalled and reinstalled the software. Any ideas as to what could be wrong? Thanks in advance.
Hi,
something to check first:
- What is the Window version ?
- Do you make sure that source code of B.04.02 you extracted from .ISO or copy from CD Rom do not get any error ?
- Do you install IO library ?
- Do you have .Net framework meet this version requirement ?
------
GiMiTEC | https://gimitec.com/catalog
------
Image
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry