What has happened when all the effluent from the detector comes out the PRG_OUT line even if Purge is set to off? Response of the detector seems to still be okay? Does anyone have a service manual for the 410? I just have the operator's manual, and purge-valve problems are not mentioned.
-
- Posts: 62
- Joined: Wed Mar 02, 2005 9:39 am
- Location: Zwijndrecht - Belgium
If the eluent is coming out of the Purge line, the reference cell of the RI detector in in the solvent flow. This means that at that point the detector cannot work. The problem you have may be in the waste/recycle line. If there is a blockage (or relative high backpressure) in this line, the eluent will flow through the purge line. Check the ID of the waste/recycle line. It should have a high ID. Using a narrow ID (e.g. 0.007" or 0.012") will cause a high backpressure on the RI detector and this flow through the reference cell.
Hope this helps.
Hope this helps.
Regards,
Gilbert Staepels
Ideas mentioned in this note represent my own and not necesseraly those of the company I work for.
Gilbert Staepels
Ideas mentioned in this note represent my own and not necesseraly those of the company I work for.
2 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science