Dear All
I am writing a investigation regarding to the continue sequential runs drop in retention time each time from one to the next, and I am wondering is there any deviance says there is any relationship between the retention time decrease vs. column deterioration?

Thank you,
As I recall the answer is no! However, there may be a requirement in the validation documentation for a RT window. If not you can always add an addendum to the AMV protocol.
Hi, fish.vegt,

I agree with HPLC Chemist in part--the method aught to include a range of acceptable retention times per peak (or relative retention times).

Is it possible that column stationary phase deterioration may cause a reduction in retention time for peaks--yes. Is this the only cause of this particular phenomenon--no.

Feel free to refer here

http://ccc.chem.pitt.edu/wipf/Web/LCMS% ... ooting.pdf

(pgs. 5-6), and

http://www.chromatographyonline.com/ret ... case-study
MattM
Thank you for the comment and the valuable reference
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: John Guajardo and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry