Can anyone provide any insight into this issue I am having please?

I run a gradient Reverse Phase method on a Waters H-Class instrument. column is C18 2.1 x 50mm, ph of sodium phosphate buffer is 3.1 and the method works very well for assay samples but periodically when I run a standard for system suit, the second or third injection has higher areas than the rest of them (for example, 5 injections of area 100000 then a 6th of 104000) and this can also happen for standards throughout the run, so in a run with 10 standard injections, one of the injection can be higher by up to 4000 and often cause the RSD % to fail specification.

Strangely, if I run the samples and standard again without using fresh standards or putting a fresh cap on the vials, the run works out fine but the random nature of this occurrence means its hard to pinpoint the cause. I use ANSI48vial 2ml vials from Waters and I don't tighten the caps too hard. Could it be something to do with pre/post inject wash? The most common time this error happens is on the 2nd injection of the system suit. Any ideas?
I wonder if a 4% error on the injection volume is outside the specs of the instrument for the volume you're injecting. It sounds reasonable. I suggest to submit a ticket to Waters, personally I have good experiences with their iRequest system.

What is the injection volume and is this fixed-loop or flow-through mode?

Is there any indication of carry-over (are solvent blanks baseline)?

An easy fix might be to run a blank or 2 before your SST.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry