I have connected waters 2489 Detector through Ehterent Cross Cable to PC and IEEE Cable from Busless to Separation Module . I have also connected Start signal cable to separation Module and 2489 Backside connector . I have installed 2489 Drives also . I am Getting Baseline in software but no peak after caffine Injection . I get MV signal on Detector rising and falling as it emerges from column . What could be Problem of Not Getting Peak in Empower II ?
-
- Posts: 13
- Joined: Thu May 31, 2018 7:54 pm
Hi:
the post you made is unclear.
If you are using an Alliance non E model then the connection to your lace box is GPIB. If you are using a 2489, the connection to the lace box is lan. You need to configure the system in Empower and make sure that the firmware of the 2489 is compatible with the ICS you are using. A compatibility chart may be available from Waters.
Run DHCP server and make sure serial # and firmware # are shown in the DHCP window, as well as in service mode of ICS(Icops.)
You do not need a trigger cable if using a 2489 with a 2695.
the post you made is unclear.
If you are using an Alliance non E model then the connection to your lace box is GPIB. If you are using a 2489, the connection to the lace box is lan. You need to configure the system in Empower and make sure that the firmware of the 2489 is compatible with the ICS you are using. A compatibility chart may be available from Waters.
Run DHCP server and make sure serial # and firmware # are shown in the DHCP window, as well as in service mode of ICS(Icops.)
You do not need a trigger cable if using a 2489 with a 2695.
-
- Posts: 13
- Joined: Thu May 31, 2018 7:54 pm
Hi:
the post you made is unclear.
If you are using an Alliance non E model then the connection to your lace box is GPIB. If you are using a 2489, the connection to the lace box is lan. You need to configure the system in Empower and make sure that the firmware of the 2489 is compatible with the ICS you are using. A compatibility chart may be available from Waters.
Run DHCP server and make sure serial # and firmware # are shown in the DHCP window, as well as in service mode of ICS(Icops.)
You do not need a trigger cable if using a 2489 with a 2695.
the post you made is unclear.
If you are using an Alliance non E model then the connection to your lace box is GPIB. If you are using a 2489, the connection to the lace box is lan. You need to configure the system in Empower and make sure that the firmware of the 2489 is compatible with the ICS you are using. A compatibility chart may be available from Waters.
Run DHCP server and make sure serial # and firmware # are shown in the DHCP window, as well as in service mode of ICS(Icops.)
You do not need a trigger cable if using a 2489 with a 2695.
-
- Posts: 4
- Joined: Mon Jun 18, 2018 12:57 pm
Thank you very much for your replay and solustion
i will try this
you have understood my post
i will try this
you have understood my post
4 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Majestic-12 [Bot] and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Majestic-12 [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science