New Mobile Phase arrangement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
Is using a mixture of 70/30 MeOH/Water from a singular bottle the same as using two separate mobile phase bottles? One containing water and one containing MeOH, set to 75% MeOH and 25% Water. HPLC-grade of course, with 4g of formic acid. I realize the amounts % are different. I am more asking about the method. I know this seems like a simple question, which is why I am asking. Sometimes the simple things are too simple and I am suspicious. I am a beginner HPLC analyst and have made much progress with help from members in this online community. Thanks!
-
- Posts: 9
- Joined: Thu Oct 26, 2017 10:49 pm
- Location: Santa Fe, NM
I recently had a question about it being possible to switch out my current mobile phase set up with a similar set up, but only drawing from one bottle. This is sort of a part B to that question.
We are testing out a different method of using 70/30 MeOH/Water from a singular bottle instead of two separate bottles. We currently run at 75% and 25% but want to do some tests to see if we can maintain efficiency and lower costs.
The column I have on now is a Raptor ARC-18 and is about 3 months used but relatively clean, maintaining pressure and retention times. I have read that it is bad practice to test out a new method on an old/moderately-used column. I have a feeling that what we are trying to do is exactly what that warning is referring to.
We are testing a sample with the current set-up and then will be changing the mobile phase to one bottle of 70/30 MeOH/Water and testing the same sample. Then we will compare results.
Should we replace the column first or can I equilibrate the used column with the new mobile-phase and then just run it as is?
Thanks in advance!
We are testing out a different method of using 70/30 MeOH/Water from a singular bottle instead of two separate bottles. We currently run at 75% and 25% but want to do some tests to see if we can maintain efficiency and lower costs.
The column I have on now is a Raptor ARC-18 and is about 3 months used but relatively clean, maintaining pressure and retention times. I have read that it is bad practice to test out a new method on an old/moderately-used column. I have a feeling that what we are trying to do is exactly what that warning is referring to.
We are testing a sample with the current set-up and then will be changing the mobile phase to one bottle of 70/30 MeOH/Water and testing the same sample. Then we will compare results.
Should we replace the column first or can I equilibrate the used column with the new mobile-phase and then just run it as is?
Thanks in advance!
-
- Posts: 517
- Joined: Thu Jul 14, 2016 12:08 pm
I'd test it on the used column.
Make sure to write down exactly how to make the MeOH/water mixture. Taking X mL MeOH and adding water to volume mark Y will not give the same mixture as doing it the other way around.
Make sure to write down exactly how to make the MeOH/water mixture. Taking X mL MeOH and adding water to volume mark Y will not give the same mixture as doing it the other way around.
-
- Posts: 9
- Joined: Thu Oct 26, 2017 10:49 pm
- Location: Santa Fe, NM
W buy our 70/30 pre-mixed. Thank you for your advice. I ran the sample today with the 70/30 mobile phase and it did not go well. Meaning the chromatography was pretty ugly. No worries though as this was just a test! Thanks again!
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
- Location: Denmark
the chromatography was pretty ugly
The description above is not valid/informative in this case.
Ugly in which way - fused peaks, or broad peaks, or unacceptable retention time............?
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
5 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: John Guajardo and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: John Guajardo and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science