-
- Posts: 172
- Joined: Thu Jul 26, 2007 2:13 pm
Peak tailing in NP chiral LC
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
I'm developing a separation for enantiomers and have a general question. I'm using an AD-H column, MP is 85:15:0.1:0.1 n-Hexane:EtOH:ESA:DEA. Compound is l and d-prolinamide. (I know it doesn't have strong UV absorbance). Anyway I obtain separation of R=4.5, but the tailing for the 2 peaks is drastically different (1.3 for L-prolinamide and 4.3 for d-prolinamide). What could cause this fro mirror image isomers? thx
-
- Posts: 172
- Joined: Thu Jul 26, 2007 2:13 pm
Turned out a MP blend was the answer. 70:22.5:7.5:.15 Hexane:EtOH:IPA:ESA was the answer. Both isomers are R=2.5, T =1.6 for both. Turns out the LC class I took several years back is still paying off. Thx Tom.
-
- Posts: 422
- Joined: Tue Sep 15, 2015 2:07 pm
The problem you experienced was that you used an acid and base (ESA:DEA) for a modifier in your method. In general, start with a neutral run, than add acid OR base, only if needed. Using DEA or other amines will result in temporary or even permanent surface changes to the column so be careful. When over-used (which we see a lot of novices do), they can result in a failure to reproduce the method.
-
- Posts: 172
- Joined: Thu Jul 26, 2007 2:13 pm
I used DEA early in the development because the author of the paper I was reading on ESA said its use along with ESA was complementary. It's not in this case.
4 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: John Guajardo and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: John Guajardo and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science