Hi,

I'm trying to analyse polar pesticides, e.g. fosetyl, phosphonic acid, perchlorate,...

I'm following QuPPE v 9.3 for sample preparation.

First experiments were conducted using a Hypercarb column. Although the column was primed was spinach extract as suggested, peaks were broad, retention times were unstable (+- 2 min) and recoveries were <10% against solvent standard and increased to 140% after correction of internal standard.
Therefore, results were not satisfying.

I switched to a zwitterionic HILIC column (Sequant ZIC HILIC, 250 x 4,6 mm, 5µm)
Recoveries increased to 60% against solvent standard and to 95% after correction of internal standard (using the same extracts as above, so this seems to be rather a high signal suppression and mathematical overcorrection issue less than an extraction efficiency issue)

Unfortunately, I do not get stable retention times on the HILIC column (+- 1 min)
I've tried:
- equilibration 20x column volumes after each injection (HILIC seems to need longer equilibration times than RPLC)
- using buffered eluents (5 mM ammoniumformiate + 0.3% formic acid, pH 3)
- prepare eluents freshly
- diluent matches starting condition
- different column temperatures

Do you have any suggestions how to overcome this retention time shifts?

Is Ion Exchange chromatography less prone to retention time shifts? I've found a paper that seems to be relevant (Anna Bauer et al, Determination of Fosetyl and Phosphonic Acid at 0.010 mg/kg Level by Ion Chromatography Tandem Mass Spectrometry)
Hi, Klamtza,

Yes, HILIC is known to require far more time for column equilibration than revere phase chromatography. This guide may help:

http://mac-mod.com/pdf/product-bulletin ... 0Guide.pdf

Chapters 1 and 8, among others. The guide is quite good, I think.

Reviewed the paper you cite above, seems reasonable work to me. Was a bit disappointed with the lack of retention time repeatability data? That said, the column used is solid, though I've not used quite that eluent with it myself. No post-column suppressor...
MattM
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry