lmh » Fri Jan 05, 2018 11:19 am
I have to disagree completely with the criticism of doing an injection with no column. Of course it's not a good way to quantify something, but there is was no suggestion that the final assay should be done without a column. It's merely a diagnostic experiment to check for a drastic disaster in the chromatography.
If you carry out chromatography on a pure standard and fail to detect anything, one quite reasonable possibility is that the analyte got lost in the column and failed to elute. Trying an injection without the column is the ideal way to test if this has happened. If you still don't manage to detect anything without a column, then it's more likely that the analyte can't be detected at the concentration you're using, with the chosen detector. Doing an injection without a column is part of the diagnostic procedure of working out why there was no peak in the original run.
Since injecting pure solvent or water can also make a peak on a non-specific detector, it makes perfect sense to compare the size of the peak in identical injections with and without the analyte, to make sure that it's the analyte that's responsible for the peak.
It's always a good idea to solve one problem at a time. If you have no idea how your analyte will behave in chromatography and no idea how to detect it, then it's hard to choose a chromatographic system that will give reasonable retention time and look for the compound simultaneously.
In terms of finding the wavelength maximum, a PDA is merely an online spectrophotometer, so it's a reasonable option for checking spectra. The only reservation is that the spectrum will depend on solvent/pH, so a spectrum determined in an injection won't necessarily be an exact match to the spectrum in a real peak from half way through a chromatogram. But it's a first good guess.
Yes, Agilent PDA will also give you a spectrum; the limitation of Agilent's PDA, at least in Chemstation, is that you can't subsequently build from the collected spectra a chromatogram at any chosen wavelength - you only get chromatograms at the fixed wavelengths you asked for in advance - frustrating, but that's how it is.
In terms of other detectors - well, yes, if you have the luxury of a range of detectors, by all means try a different one if the analyte isn't likely to have good absorbance at a nice long wavelength, but not everyone has CAD and MS available, and if PDA's all you've got, and the analyte has at least some absorbance, it might be adequate to use this detector even though in an ideal world you'd choose something else.