Waters 410 RI sinusoidal baseline problems
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
I am using the 410 with a Hitachi D-7000 LC system. The 410 is exhibiting a sine wave in the baseline, with a period of about 2 hours. When I switch into purge, the sine wave has a higher amplitude and shorter period. Anyone seen something like this? Welcome to ideas.
-
- Posts: 198
- Joined: Tue Mar 27, 2007 6:54 pm
- Location: Belgium
The sine wave is generally common for RI and I have always thought that it was due to the heating cycles of the column oven. In your case it seems more the heating cycle of your RI (at what T are you using it?).
Maybe you can switch off the column oven and see what happens?
If the temperature of your lab doesn't change extremely fast you should get a drifting baseline (either upwards or downwards) after a while..
Normally the sine baseline is not so critical if the signal is high enough and the amplitude is reasonable (at least considering I use RI only for GPC).
Maybe you can switch off the column oven and see what happens?
If the temperature of your lab doesn't change extremely fast you should get a drifting baseline (either upwards or downwards) after a while..
Normally the sine baseline is not so critical if the signal is high enough and the amplitude is reasonable (at least considering I use RI only for GPC).
-
- Posts: 6
- Joined: Fri May 11, 2018 1:01 am
Looks like a Temperaure fluctuation, what temperatures are you using in your column oven and inside temperature of your Waters 410?
-
- Posts: 415
- Joined: Tue Sep 15, 2015 2:07 pm
That sounds like temperature fluctuation for sure. Here is an article link that will help you understand and troubleshoot the problem.
"HPLC Baseline Stabilization Tips for Refractive Index Detectors (RI or RID)" ;https://hplctips.blogspot.com/2018/02/hplc-baseline-stabilization-tips-for.html
"HPLC Baseline Stabilization Tips for Refractive Index Detectors (RI or RID)" ;https://hplctips.blogspot.com/2018/02/hplc-baseline-stabilization-tips-for.html
4 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science