Looking for suggestions on silica/silicate analysis without the use of ICP or post-column derivitization for supressed conductivity. Any thoughts?

Direct (unsuppressed) conductivity should work with a high-pH buffer. The conductivity signal will decrease when silicate elutes.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
I've had some luck in the past with an Ion Exclusion column using just water as the mobile phase and conductivity as the detector. Due to the low background the method is also quite sensitive. The Dionex ICE AS1 column worked well for me. We were using this for high purity water applications, i.e. there were very few other anions in the matrix. If you have alot of other things in the aqueous matrix then this would not be a good choice.
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 1 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry