Can anybody suggest me good C18 column which has long life , because I have to purchase it now.Pls suggest me.

This is kind of like taking a shot in the dark because each C18 column on the market has unique properties .... But:

Waters Symmetry C18
Phenomenex Luna C18(2)
Agilent Zorbax SB or Eclipse C18's
Waters XTerra C18 (very rugged, wide pH range, unique selectivity)
Phenomenex Gemini (actually a C14, very rugged, wide pH range, unique selectivity)
Can anybody suggest me good C18 column which has long life , because I have to purchase it now.Pls suggest me.
Zorbax Eclipse XDB from Agilent or YMC-Pack Pro C18 (1-10 pH) - both has excellent peak shape and long life

Phenomenex Prodigy, GL-Science's Inertsil, Zorbax SB, Waters Symmetry, Thermo... all make excellent columns, as do several other manufacturers. Your success will depend, somewhat on how well your analytes match the packing you choose.

If you have a basic narcotic (for example), your goal should be to pick a column that will minimize tailing (Symmetry or other highly endcapped column w/ ultra pure Si, minimal silanoles - look at their Amitriptyline peak shapes).

If tailing isn't a concern, look at columns that show good stability over thousands of injections (as I've done w/ Prodigy columns).

If you're going to wind up having lots of labs doing the same test on the same column, lot to lot reproducibility may be your key to success (Inertsil packing is good for this).

Somewhere among these criteria is a (hopefully large) handful of packings that will work well for you.

If you are certain that longevity is your main concern, I'd suggest Prodigy, Stable Bond and Inertsil C-18 packings based on my own experience w/ clean pharma samples and these packings.
Thanks,
DR
Image

It has been mentioned before already, but the best choice depends on what you want to do with the column. How long a column lasts, will also depend on how it is protected (see the recent discussion on guard columns).

Best choice for low pH work (<2): Atlantis T3
Best choice for high pH work (9-10 or higher): XBridge C18
Best choice for peak shape: a packing with an embedded polar group such as SymmetryShield RP18.


You did not ask for this, but in some environments the column is used for the same assay over and over again.Therefore: best choice for long-term reproducibility: Symmetry C18 (or C8).
In the Addition of Mr. Uwe Neue . Dr. , shaun78
the Best C-18 Column is Depends on Your Requirment Like Sample Loading and Proper Washing and Storage of Column after work hasbeen Completed. if You purchesed any Brand Name Column and Proper care is not . Sample Loading is Very High then any column cannot be stable more then 400 to 500 injection.
So Please Be Aware About Sample Loading , pH Range , Solvents which are not Suitable for Those Particular Column.

What type of compounds will you be working with?
i am very glad on receiving the suggestions from all of you.

We had a situation a year ago where the column we used was not manufactured anymore. We had to change and I did a small study with various columns to see what we should go for. The analysis was a simple isocratic assay method with a phosphate buffer pH 7 / ACN mobile phase. The product was a peptide including some PVP and other excipients.

About 100 injections on each of the following columns:

Phenomenex Gemini
HiChrom ACE C18
XTerra C18
Zorbax Eclipse C18

The results I got was that the Ace column gave the best plate count and peak symmetry - in the beginning. The column deteriorated for every injection we did. The Phenomenex was worst in all aspects, bad tailing from the start and fading over time.

We ended up chosing the XTerra column since it was the most stable. It had actually the lowest plate count of all columns, but it didn't change. The symmetry was 1.2 for all 100 injections. Since it was to be used in the production with many thousands of samples per year - stability was the primary concern.

This study is of course only valid for our product, but it can be a hint anyway.

When evaluatting C18 phases, one should look at the following data from vendors:

pH durability
Selectivity (hydrophobicity, hydrogen bonding, steric selectivity)
Injection repeatability - both for isocratic and gradient (more challenging)
Residual Silanol Activity (basic compound at neutral pH)
Column Efficiency
Column Back Pressure (since HPLC instruments have pressure limits)

You can still do a lot of evaluation based on vendor data alone,
but I agree that it's important to put it to the test on your own compounds.
"a C18 is not a C18 is not a C18" :)

In the early days of HPLC, octadecylsilane (most often referred to as an ODS or C18 phase) was one of the first bonded stationary phases that became available for the new technique called "reversed-phase chromatography." It became the standard phase for reversed-phase chromatography and quickly was adopted by most practitioners. Because the pharmaceutical industry was an early adopter of HPLC and regulatory bodies did not want to bless a particular manufacturer's brand of column, the FDA and USP developed a classification system that gave a generic designation for each new method that was submitted under a new drug application. For HPLC columns, an "L" designation was given and because C18 was used in a majority of the submittals, it became "L1." As additional phases were added, they were given their own "L" number (for example, L7 C8, L10 CN, L11 Phenyl, and so forth).
Unfortunately, this designation system proved to be unreliable because each commercial C18 column was synthesized differently using silica gel as the base material and the resulting reversed-phase columns had different properties. For example, some manufacturers used octa-decylmonochlorosilane and a low surface area silica gel (Figure 1) while others used the same silane but bonded it to a higher surface area silica gel. These two C18 columns would behave differently with the latter having more C18 phase than the former. A lower coverage bonded phase sometimes had unreacted silanols that caused mixed retention mechanisms. Some manufacturers use di- and trichlorosilanes and polymerized the bonded phase to form a thicker layer with different diffusional properties. To cut down on unreacted silanols that remained after the bonding process, some manufacturers endcap these silanols with a small silane (for example, trimethylmonochlorosilane). Silanols are sometimes responsible for tailing of basic compounds at intermediate pH values. Some manufacturers even went further and double-endcapped with a second small silane to provide a more inert surface. Some manufacturers used polymeric base materials and then bonded C18 moieties to their surface resulting in a totally different C18 packing material, yet it was still considered an "L1." Others use organoalkoxysilanes that display a different degree of reactivity and can produce a slightly different C18 bonded phase than would an organochlorosilane reagent.

So, the statement "a C18 is not a C18 is not a C18" holds true even today. All C18 phases are not alike and the user should make sure that once a method is developed, he or she sticks with the same column part number to ensure rugged and robust methods. In recent times, Dolan and Snyder (1,2) have studied a large number of C18 phases in an attempt to determine which phases have similar retention and selectivity features. By characterizing the phases based upon probes with as many as five retention parameters, they have classified phases with a numerical value so that one might be able to choose the closest phase in performance characteristics in case the original phase is not available from a particular manufacturer. Their approach also allows one to find a C18 phase totally unlike their existing phase that can aid in column selection during method development.
you may want to check the following link for a comparison of C18 columns. I refer to this often to choose/ change a column during method development.

http://www.ace-hplc.com/assets/download ... e_2006.pdf
Take care

I would be very careful to use such a catalogue, since it is advertising material for HiChrom Ace. You can find very similiar catalogues from the other vendors showing the superiority of their columns ;-)

:D I have never used an ACE column except for Cyano ones (and they are pretty good actually).
A similar list is available from http://www.lifescience.ca/pdf/870.pdf
It is more about column test mixture. I hope this looks unbiased to you Mattias.
The reason I refer to the literature from Hichrome website is because it lists out the features of a lot of other columns, except their's and it helps me choose a column for a particular application.
Take care
14 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Bing [Bot], Google [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry