I assume this is a new problem, the analysis worked fine before, the peaks are at the same retention time, are sharp, and are not carry over from sample injections, or new types of samples.
I also assume that when you say " purge at 3 mls/min ", you are flushing the column, and not just purging the degasser. 3 ml/min sounds rather high for flushing modern instruments with water methanol mixes, given the relatively high viscosity changes that occur.
What happens with a sequence of just multiple blanks - do the peaks sizes get smaller with each injection, or stay the same size?.
If you are running a gradient, what happens if you extend the final time by 15 minutes and then run a blank. Do the peaks change in size or retention?.
If the assumptions are OK, the most likely causes are:-
One of your reagents is contaminated - assuming you are using the same grades and sources as before, and haven't opened a new bottle of formic acid, I'd first look at the water, then the methanol, then the formic.
You can test each major component out by running 99:1 of A:B for 15, 30, 60 mins, and performing two blank runs after each time period. Repeat the process with 1:99 of A:B for 15, 30, 60 mins. If the longer periods show bigger peaks, you've found your source.
You can check the formic by halving the concentration in both A and B, and seeing if the peaks get smaller.
Peaks in the second blank should significantly decrease in size. If not, you may have contamination of the blank, the column, or the injector. You may be able to evaluate those by changing the size of your blank injection volume, and using a different source of blank solvent.
Another possibility ( of many ) is that you samples are dropping out of solution in the injector or guard, so ensure your samples are dissolved in the initial mobile phase.
I'd flush the column at the normal flow rate, and I slightly disagree with Philippe's comment about flushing the column with only the strongest solvent - because that assumes all the impurities behave similar to the analyte.
If you start seeing peaks where they shouldn't be, then either change the gradient, or flush with solvents that remove them. In some cases, a high % water flush may be needed to clean out unwanted material that deposits on the column or injector. If it's a regular problem, you should modify the method.
Please keep having fun,
Bruce Hamilton