Just a question regarding pH control in HILIC.

If you are running an isocratic HILIC method with a very high organic content (approx. 95%) is the pH of the buffered aqueous mobile phase as important (in terms of analyte ionisation) as in reverse-phase, and how efficiently does the buffer control pH when it is present in such a small percentage within the mobile phase?

Thanks.

Probably less of a problem than you might think, because the high % organic will keep your analyte from ionizing. For a thorough discussion of the whole topic, see these LC-GC articles by Bill Tindall:
http://www.lcgcmag.com/lcgc/data/articl ... rticle.pdf
http://www.lcgcmag.com/lcgc/data/articl ... rticle.pdf
http://www.lcgcmag.com/lcgc/data/articl ... rticle.pdf
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 19 users online :: 1 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry