Hi all,

We are analyzing gentamicin by HPLC-ECD according to European Pharmacopoeia. And in the method they say use dioxide carbon free water for preparing mobile phase.
How could I prepare such the water. Is it OK if I use just DI water or bottled water from Merk.
Thanks for any suggestion.
The EP probably has directions to prepare CO2 free water. This is from the USP:

Water, Carbon Dioxide-Free—It is Purified Water that has been boiled vigorously for 5 min or more and allowed to cool while protected from absorption of carbon dioxide (CO2) from the atmosphere, or Purified Water that has a resistivity of NLT 18 Mohm-cm.
scottythree wrote:
The EP probably has directions to prepare CO2 free water. This is from the USP:

Water, Carbon Dioxide-Free—It is Purified Water that has been boiled vigorously for 5 min or more and allowed to cool while protected from absorption of carbon dioxide (CO2) from the atmosphere, or Purified Water that has a resistivity of NLT 18 Mohm-cm.

Thank you very much !
That said, I suspect that He sparging would get you there too. Once left exposed to air, your water will start to re-equilibrate to about 3ppm CO2.
Thanks,
DR
Image
thohry wrote:
Is it OK if I use just DI water or bottled water from Merk.


Electrochemical detection is sensitive to a lot of organic compounds. I therefore would not use DI water as it still can contain organics.
I recommend to use ultrpure water out of a purification system applying UV digestion (Millipore, Barnsted ELGA etc.).
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry