I´m trying to analyze Clindamycin palmitate hydrochloride in an oral powder by rp-hplc. My problem is that it has a lot of tailing. I´ve tried lots of methods and always the same. Tried with C18, C8, C6, TMS and CN, different pHs but couldn´t find a solution. It retains a lot. Does anyone know a good method to try? Perhaps Normal Phase or HILIC?
Q. F. Ignacio Viera
Most of the Google links to Clindamycin palmitate HPLC are over 30 years old and use ion-pair with RI detection. What kind of detector were you using and is there a chance that you are overloading the column to try to get a detectable peak?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanks for your reply Tom.
I´m injecting 10-20 µL. I suspect column is not overloaded. I´m using DAD.
Q. F. Ignacio Viera
Recently I was trying to react metabisulfite with benzaldehyde to form an aldehyde-bisulfite adduct. I can see benzaldehyde, but not the adduct. And I think reaction is happening because benzaldehyde peak gets lower. Perhaps I cannot see adduct by UV. Fluorescence? Which lambda emm and exc?
Q. F. Ignacio Viera
I´m injecting 10-20 µL. I suspect column is not overloaded. I´m using DAD.
As far as overload is concerned, what matters is not the injection volume so much as the mass injected.

Saying you used a DAD doesn't tell much. What wavelength. If you look at the structure of clindamycin, there is no much in the way of a UV chromophore, which means that you will probably be limited to the "end absorbance" of carbonyls down around 210 nm -- and probably won't have a very high molar absorbance even there.

All of which makes me suspect that if you are injecting enough material to see a peak, you may well be overloading your column.

If this were my problem, the first thing I would do is to obtain a UV spectrum of clindamycin in something approximating your mobile phase. That will let you establish the absorbance maximum and at least estimate the molar absorbance. Then go and do a loading study -- inject your standard, then dilute by a factor of 5 or so and reinject. If the peak shape improves, then you were overloading. If it doesn't improve, that will confirm that overload is not an issue.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry