Quantitative question
Posted: Wed Jun 21, 2017 3:19 pm
by resolution1
When monitoring peaks at 215 or 220nm (peptide bond) and comparing 2 different proteins. If both are loaded at the same conc (1g/l, same volume injection), but differ in molecular weight by 2, would you expect the peak area/height to be the same or since one has double the molar mass would it give twice the area at 220nm?
Re: Quantitative question
Posted: Wed Jun 21, 2017 4:19 pm
by HPLC chemist
Nope! Any molecule including peptides and proteins can have a dramatically different response to a detector. This can be especially true for a protein that has only a small but significant change (perhaps position) in an amino acid. Therefore, for each molecule I would develop a best fit (probably linear) curve of concentration versus detector response. In other words 'PROVE YOUR ASSUMPTION'!
Posted: Wed Jun 21, 2017 6:20 pm
by Rndirk
There is no relationship between absorption of UV light and molar mass.
You can however assume that, for the same
molar concentration, a molecule with a larger amount functional groups like alcohols, phenyls, halogens, amines, double/triple bonds,.... will have a higher UV-Vis absorption in general (200-800nm).
Posted: Wed Jun 21, 2017 11:56 pm
by DR
^ until you have variations in folding for proteins... folding differences can compound the vagaries of molar absorptivity too.
Posted: Thu Jun 22, 2017 3:44 am
by dap
It is impossible just from scratch answer such questions if analytes aren't simple homologues. F.e. even benzene, benzyl alcohol, benzaldehyde and benzoic acid have similar in general but completely different UV spectra. There is quantum-chemical software to answer such questions in days of computer work.