- Posts: 2
- Joined: Wed Mar 28, 2012 2:46 pm
Linearity issues
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
We are currently analyzing phenol in UPLC system at 240 nm. Linearity of phenol (0.1 mg/mL -0.7mg/ml)is achieved in UV detector using in dual wavelength mode but the same cannot be achieved with PDA detector . Any thoughts
-
- Posts: 12
- Joined: Fri Mar 04, 2016 7:08 pm
You need to provide _full_ details of your analysis. Are the measurements carried out on on the same HPLC instrument with two detectors, or is it completely separate systems? And why measure at 240 nm? What is the mobile phase/matrix? λmax for phenol should be ~270 nm, did you determine it experimentally for your solvent system/samples?
-
- Posts: 5329
- Joined: Thu Oct 13, 2005 2:29 pm
- Location: Maun Botswana
raman_srividya wrote:
We are currently analyzing phenol in UPLC system at 240 nm. Linearity of phenol (0.1 mg/mL -0.7mg/ml)is achieved in UV detector using in dual wavelength mode but the same cannot be achieved with PDA detector . Any thoughts
Details, details, details. We can't help without the details. How may levels of standards, how many replicates at each level, and how many times have you repeated the calibration ? Also the details of the analysis. The ore you don't tell us the more we can't help you. Why are you calibrating over such a narrow range ?.
Peter
Peter Apps
3 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science