- Posts: 6
- Joined: Thu Oct 13, 2011 4:34 pm
Run 1 Run 2
705664 604274
705246 610078
702745 613880
701808 618958
699750 629232
The issue does not seem to be carryover, as there is nothing detectable in the blank injections immediately following any sample or standard injections. A brief summary of the method is outlined below.
EDTA derviatized with ferric nitrate nonahydrate solution
Buffer/Mobile Phase A is tetrabutyl ammonium bromide and sodium formate in water
Mobile Phase B is Methanol
EDTA elutes during 98% MPA and 2% MPB
The column is a Hypersil BDS C18 5 micron 4.6 x 250 mm
UV detection at 260 nm
Any insight would be greatly appreciated.