I routinely run SEC samples on a silica gel-based high efficiency SEC column and was happy with it until I encountered non-specific interactions between my analyte and the stationary phase in one series of my samples. The analyte was a polymeric nanoparticle with a hydrophobic small molecule encapsulated in it. I ran it in an aqueous buffer and expected to see a nice peak corresponding to the size of the polymer particle. To my surprise, the SEC phase acted more like a reverse phase retaining the small molecule and letting it elute way beyond the limit of the small molecules for this SEC column. The peak attributable to the particle was there, but it was an empty particle. Adding 10% methanol moved the peak of the small molecule to an earlier time, but still way beyond the limit for small molecules, adding a piece of evidence to support my hypothesis. Tech support for the column confirmed that if I have a very hydrophobic molecule, it can be retained non-specifically on the silica-based stationary phase.
I noticed that there are polymer-based SEC columns available on the market. Can anyone recommend a good brand? Also, do you know if they are much more hydrophilic than silica-based ones? I have good relationship with people at the company which sold me the column I am currently using, but, apparently, they are not making polymer-based columns and did not want to comment on them.
Here is what I found so far:
http://www.pss-polymer.com/products/col ... c/suprema/. But I do not know the brand and am not comfortable about choosing the column without knowing enough about the phases.