Dear all,

I have the challenge to set up a methods for the analysis of chondroitin sulphate and glucosamine by HPLC-UV or HPLC-MS/MS... I know there are methods on the web that used IR and Light scattering detectors but I don't have them.
Any comment or suggestion will be much appreciate.

Looking forward to hearing from any of you,

Carol
Dear Carol

You can do the analisys of glucosamine by HPLC but the chondroitin is not possible.
Look at the USP 29 online from China.

Saludos (Greetings)

Fernando :)
http://www.sielc.com/Compound-Glucosamine.html

You need to use ELSD for detection.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Thanks for the replay,

But I am sure that they should be a way of doing it,
Dear Carol

I found in greenmtlabs.com that by enzymatic hplc that treating chondroitin with chondroitinase ABC they produce disaccharides that can be monitored at 240 nm with rphplc and are very neatly separated.

Buena suerte (Good Luck).

Fernando :(
Chondroitin is not UV active, in order to use UV you need to do derivatization or indirect detection. Any UV active impurity in the sample will mislead you.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Shouldn't each of these be detectable using positive mode MS due to the presence of amine functional groups? I suppose the separation could be difficult because of the large difference in MW but it could be worth a try.
7 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry